Good Deal! Only one issue on my part...It's too small. I had to have the HP3
Did you hook 12 VDC to it? without the power cord, it will not read PEP correctly.
What does it show with a carrier?
I made a call to the people that handle Diawa in usa they say it could 20% off at half scale and 10% off at full scale. I do not consider that acceptable.
I am not totally happy. I am using a dummy load. More testing this weekend. Then it may go back.
If you would like to know more on the matter you can look at ANSI/NCSL Z540-1-1994. That is the spec that calibration labs have to follow. You can use a NIST STANDARD or transfer standard for your instrument calibrations...either way there is a 4:1 test accuracy ratio simply meaning whatever instrument you are calibrating has to be done by another instrument that is 4 times more accurate. And talk about compounding percentages that spec also says that the collective uncertainty of the measuring standards shall not exceed 25% of acceptable tolerance for each characteristic. lol.